    The electrode should be clean to start with, and the plating should be done under dim light. After all, you are forming a silver halide. It is light sensitive.

    PE

  2. #22
    kb3lms's Avatar
    Join Date
    Jun 2006
    Location
    Reading, PA USA
    Shooter
    35mm
    Posts
    697
    Images
    5
    Some better results!

    Today, I re-plated the RE and the Bromide electrode under safelight. Now the 0.1M KBr solution gives a reading of -68.0 t -68.5 mV. Ambient temperature was 20C so according to the chart, that should give a reading of about -62.56 mV. Since my solution has sat for a few days there may have been some evaporation and it could now be at higher concentration and/or it could have been a bit off in the first place. Later this evening I will make fresh solutions and get new readings as Kirk suggests above.

    -- Jason
    All this has happened before, and all this will happen again.

  3. #23
    Photo Engineer's Avatar
    Join Date
    Apr 2005
    Location
    Rochester, NY
    Shooter
    Multi Format
    Posts
    23,027
    Images
    65
    Jason;

    You are as close as I get sometimes, with an offset of about 6 mv.

    You should only apply voltage (plate) the Silver electrode, and you do it with another Silver electrode in the halide you wish to plate it with.

    PE

  4. #24
    kb3lms's Avatar
    Join Date
    Jun 2006
    Location
    Reading, PA USA
    Shooter
    35mm
    Posts
    697
    Images
    5
    PE,

    Yes, sorry. I meant that I re-applied plating to silver electrode in KBr (.17M) and also replated the reference electrode in KCl (0.05M). My reference electrode was built from the instructions in the article I posted above. It seems to work quite well. Compared with the RE in my Hanna pH meter, the readings are about the same.

    Also, just for the record, my silver wire is 14 gauge .9999 fine silver bought off the big auction site. Quite sure the real seller is one of the references given in other posts here. 14 gauge is sturdy enough to use for the electrodes.

    -- Jason
    Last edited by kb3lms; 10-13-2012 at 07:41 PM. Click to view previous post history.
    All this has happened before, and all this will happen again.

  5. #25
    kb3lms's Avatar
    Join Date
    Jun 2006
    Location
    Reading, PA USA
    Shooter
    35mm
    Posts
    697
    Images
    5
    Also, for anyone following this thread, here are instructions (courtesy of RIT) for plating the silver electrode using a piece of equipment they have, or more likely had. All it really did was run the plating process for a fixed four minutes. There is a description in one of the papers I found giving some description of the machine.

    https://www.dropbox.com/s/w50e1hm3a56hvri/plating.pdf

    Also, while their method of cleaning the electrode does work, I am just cleaning mine with scothbrite. Works great and it gets nice and shiny. Initially I did sand down the sharp edge left by the wire cutters but have not sanded since.
    All this has happened before, and all this will happen again.

  6. #26
    Photo Engineer's Avatar
    Join Date
    Apr 2005
    Location
    Rochester, NY
    Shooter
    Multi Format
    Posts
    23,027
    Images
    65
    Well, the restriction on the length of exposed plated Silver is not a hard and fast rule.

    Also, the HANNA meter electrode is not ideal for this measurement. My RE is a HANNA Ag/Ag/Cl electrode build as a RE and it works fine.

    PE

  7. #27

    Join Date
    Jun 2004
    Location
    Portland, OR
    Shooter
    4x5 Format
    Posts
    3,268
    I used to perform Total Organic Halide analysis, and we had a silver electrode in it that we would clean using Ajax Powder, which has chlorine in it, well, probably as chloramine-T which releases free chlorine in water solution. The Ajax was nice as it had fine grit for scrubbing the crap off the silver electrode and I'm sure plated it somewhat as well!
    Kirk

    For up from the ashes, up from the ashes, grow the roses of success!