Measure the resistance of both of the heaters at the terminals on the ends of the wires with one end disconnected. 185 ohms sounds kind of high. Also measure the voltage on the terminals the heater is connected to. Without knowing the supply voltage, it is hard to determine a reasonable range for the resistance of the heating elements. Careful when doing this, you don't want to make any sparks. 
The thermal paste would thin when heated, it is a suspension of zinc oxide in a light grease and if over applied it would drip off.
-
 Originally Posted by Bob-D659
Measure the resistance of both of the heaters at the terminals on the ends of the wires with one end disconnected. 185 ohms sounds kind of high. Also measure the voltage on the terminals the heater is connected to. Without knowing the supply voltage, it is hard to determine a reasonable range for the resistance of the heating elements. Careful when doing this, you don't want to make any sparks.
The thermal paste would thin when heated, it is a suspension of zinc oxide in a light grease and if over applied it would drip off.
I'm on UK mains voltage which is around 240V. The ACP505 does however contain a small transformer which I assume is used for the control circuitry.
Tom.
-
Ok, yes the thermal paste might drip off when over applied, it is probably just a suspension of zinc oxide in a light grease. That resistance measurement of 185 ohms of the heating element sounds way high. Try measuring again at the terminals at the end of the braid covered wires with one of the heater terminals disconnected. Also check to make sure the crimp connection to the wire looks good. You should also measure the supply voltage to the heaters when the heater is disconnected. That is needed to determine what range the resistance of the heating element should be in. For example, if the supply voltage is 12 volts, the element would only draw very little current and produce no heat at all. Careful here, don't short anything out.
Sorry about the sort of double post. If those are mains voltage heaters, that would explain the high resistance. Be more careful even.
Last edited by Bob-D659; 05-22-2009 at 08:59 AM. Click to view previous post history.
-
Bob,
Each of the inline heaters are 230V / 250W spec. I wonder if an overly high mains supply voltage (say 250V) could cause the heaters to "burn out"...
Tom.
-
Well that resistance is about right for a 230v/250 watt rated heater, so I doubt they are burnt out. I'd trace the circuit back and see if some other component is "open", possibly like a fuse on a circuit board, or maybe that silver coloured controller.
-
Sponsored Ad. (Subscribers to APUG have the option to remove this ad.)
-
acp 505
Hi Tom
I have been working with these machines for many years.
From what has been described, the heating problem does not seem to be the heat exchangers or the thermal cut outs. I would recommend testing the thermostat, the two mechanical relays on the thermostat board and check the ciculation!
Regards
Paul
-
Paul,
Thanks for your post. Did you repair the machines at one point in time? The circulation is working, you can see (and feel) the solutions (or plain water in the test case) coursing through the system.
Tom.
-
-
Tom, the last device is an adjustable resistor or potentiometer. But in contrast to the more common ones it is not of the radial but of the axial type.
The use of a threaded spindle allows very precice adjustments. It is called Trimpot.
-
|
|